LCGUY » Fri Aug 31, 2007 6:21 am
Hi Rick,
As everybody explained above, the main reason of loss of separation or RT shifts with Reverse phase columns (specially C18) is
dewetting of pores or matting of carbon chain.
I believe these two are inherently different phemonenon (debatable subject for chromatographers). But reverse phase columns with some polarity introduced are unlikely to pose problems in your results.
I used PrincetonSPHER Ultima C18 and C8 from Princeton Chromatography Inc. (this has poler amide group embedded in alkyl chain) for typical applications where my MP is 100% aqueous and I got excellent results with no such effects like loss of retention..blah.blah.
Princeton has one more typical phase which has Y shaped C18 carbon chains for 100% aqueous mobile phases. You have 4 carbons and then bifurcated chains of 7 carbons each..something like that.... This is based on the principle that Y shaped chains prevent each other from collapsing under highly polar (aqueous) mobile phase environment. Surprisingly this column also work under 100% water conditions. WHERE DOES
DEWETTING THEORY STANDS IN THIS CASE??? BUT THIS EXPLAINS PHASE MATTING EFFECT UNDER AQUEOUS CONDITIONS WELL.
Here are two article from LCGC which expplains these theories in simple and detailed way.
1. The Retention Behaviour of Reversed-Phase HPLC Columns When Used Under 100% Aqueous Mobile-Phase Conditions
http://www.lcgceurope.com/lcgceurope/da ... rticle.pdf
2. Phase Collapse in Reversed-Phase LC
http://www.lcgceurope.com/lcgceurope/da ... rticle.pdf